GenesForLife wrote:Carried out conductometric titrations today with 0.1N HCl and 1N NaOH , the HCl used came with a composition of 1.18 g/ml at 20'C, I prepared the solution at 35'C , data from two sets of titration resulted in a normality estimate of 0.04N , less than half the expected concentration, what could have gone wrong?


GenesForLife wrote:the HCl used came with a composition of 1.18 g/ml at 20'C


Cito di Pense wrote:GenesForLife wrote:the HCl used came with a composition of 1.18 g/ml at 20'C
That's a density, not a composition, starshine. But presumably, HCl has a density-composition curve at 20°C and 1 bar. If that yields 0.1 N, then that's what it yields. But you would know better than I would, especially if you prepared the solution at 35°C.
How did you calibrate the conductimetry?

GenesForLife wrote:Cito di Pense wrote:GenesForLife wrote:the HCl used came with a composition of 1.18 g/ml at 20'C
That's a density, not a composition, starshine. But presumably, HCl has a density-composition curve at 20°C and 1 bar. If that yields 0.1 N, then that's what it yields. But you would know better than I would, especially if you prepared the solution at 35°C.
How did you calibrate the conductimetry?
Conductometer has built in calibration.
I think it can still be used to deduce required volume for desired mass when working with normalities.
Density = mass / volume reduces to mass/unit volume, therefore required mass can be calculated as a function of the volume.


GenesForLife wrote:Precisely, which is why we had results at half the expected value, never mind, I've solved the problem now, I found what I was looking for, a standardized table for temperature-density relationships.
http://www.handymath.com/cgi-bin/hcltbl ... bmit=Entry ,



GenesForLife wrote:But on second look, things don't add up, the density (and therefore resulting normality) increases with temperature (according to that calculator) , in which case our lab temperature should have resulted in stronger HCl and not weaker HCl, I need to look elsewhere.


GenesForLife wrote:Nope, you are actually correct in noticing that temperature can cause changes in density,and therefore affect the normality of any solutions prepared on that basis, when using liquid solutions, the norm is to convert normalities to volumes ( If at 20' C, I needed 1.18 grams of HCl to make my solution, I'd take 1ml of the stock solution because it contains that much mass, and so on) , if the density differs, I'd get a different quantity of HCl in the same volume of solution, thus resulting in a different (and wrong) value for normality.


GenesForLife wrote:Our protocols don't account for temperature variations during calibration, do you know of any that do?



GenesForLife wrote:Yup, that is correct, but the solution I prepared took that into account as I took a volume that contained the weight/mass of HCl I needed.